Analysis of C16 to C24, which column?

PostPosted: Wed 22 Jun 2016 09:44
by Miella
Hello,

I make the analysis of C16 to C24.
I use a DBWAX and the results are good but my columns resist badly with methanol and sulphuric acid.
Do you have an other solution to make this analysis?

Thank you,
Miella

Re: Analysis of C16 to C24, which column?

PostPosted: Wed 22 Jun 2016 09:55
by Nils
Hi Miella,

Sulfurid acid in general is a problem for GC capillary columns, even for crosslinked and bonded phases, independent of the polarity of the column, WAX or DB-1. The best recommendation is to get rid of sulfuric acid prior to analysis or to accept that the lifetime is not as good as without.
Sorry

Nils

Re: Analysis of C16 to C24, which column?

PostPosted: Wed 22 Jun 2016 10:54
by Miella
Thank you Nils !