CALIBRATION AND PERFORMANCE TRACKING OF FLOW CYTOMETERS USING SPHERO™ CALIBRATION PARTICLES
SPHERO™ - Spherotech, Inc.


Introduction
The SPHERO™ Calibration Particles are versatile, stable, economical and convenient to use. These particles contain a mixture of fluorochrome which are spectrally similar to many of the fluorochromes used in flow cytometry. As a result, they are used for routine alignment, day-to-day performance verification and long term performance tracking of several channels of flow cytometers in one run. These particles are very stable since the fluorochromes are entrapped inside the particles instead of being located on the surface. They are packaged in a convenient dropper bottle to facilitate the dispensing and storage. The diluted particles can be stored for later use if desired to reduce costs.
These products and their uses are described briefly as follows:

Rainbow Calibration
The RCPs are designed for the routine calibration of most available channels in any flow cytometer. For example, these particles are used to verify the instrument set up and to check the linearity and sensitivity of the instrument. If factory recommended procedures are used for instrument set up, we recommend that the RCPs are included into QC programs to track the long term and day-to-day performance.

The RCPs contain a mixture of similar size particles with different fluorescence intensities. Each particle has a mixture of fluorescent dyes entrapped inside. This allows the excitation and detection of the particles in most channels of any flow cytometer. For example, Figure 1 shows the histograms of the Spherotech Cat. No. RCP-30-5 in the FITC, PE, ECD, PE-Cy5, and APC channels.
Image
In addition, the MEF (Molecules of Equivalent Fluorochrome) values of each peak in the RCP-30-5 have been assigned for the FITC, PE, ECD, PE-Cy5, and APC channels.

Since MEF values do not specify the fluorophore used or the intended channel in flow cytometer, Spherotech has decided to use more specific terms, namely:
MEFL (Molecules of Equivalent Fluorescein), MEPE (Molecules of Equivalent PE) and MEPCY ( Molecules of Equivalent RPE-Cy5), etc. However, the users are welcome to use whatever terms they prefer.

After obtaining the histograms, the Relative Channel Number of the RCPs vs. the MEF (ie. MEFL, MEPE or MEPCY) is plotted to obtain Calibration Graph in
all channels as shown in Figure 2.
Image
These graphs are used to determine the linearity of the instrument and corresponding MEF values of stained cell samples
according to the procedures described in Spherotech STN-9. The collection of the obtained data provides day-to-day performance verification. This data can be collected over several years for long term performance tracking of the instrument.

Ultra Rainbow Calibration Particles (URCPs):
New flow cytometers with an increasing number of fluorescent channels are now available. As a result, the Ultra Rainbow Calibration Particles with enhanced UV and Far Red fluorescence intensity have been designed for performance tracking of flow cytometers with multiple lasers. For example, the Ultra Rainbow Calibration Particles have excellent resolution in the
UV, violet, green, yellow, orange, red, far red, and IR channels of the flow cytometer. The URCPs are available in 6 peaks, 3.8μm or 5.1μm.

These particles are similar to the RCPs since a calibration graph for the URCPs is obtained using the same
procedure. For example, Figure 3 shows the histograms for the URCP-38-2K, while Figure 4 shows the calibration curves. However, the URCP have a broader excitation wavelength range and improved resoltuion in the UV and Far Red Channels.
Image
Image

Rainbow Fluorescent Particles (RFPs)
The Rainbow Fluorescent Particles (RFPs) contain uniform size particles with a single intensity. The single peak of the RFP has low fluorescence and size coefficient of variation (CV). As a result, they are useful in the alignment of the optical system of the flow cytometer in multiple channels. For example, Spherotech Catalog Number RFP-30-5 is used to align the FITC, PE, PETR, PE-Cy5, and APC channels of the flow cytometer. See Figure 5 for the histograms of Spherotech Cat. No.
RFP-30-5.
Image

The RFPs contain similar fluorophores to the Rainbow Calibration Particles. In addition, the RFPs have similar fluorescence intensities to the brightest peak of the corresponding Rainbow Calibration Particles with the exception of RFP-50-5, FP-60-5, RFP-100-2 and RFP-30-5A. The RFP-30-5A has the fluorescence intensity similar to stained cells in all channels.

Ultra Rainbow Fluorescent Particles (URFPs)
The Ultra Rainbow Fluorescent Particles are similar to the Rainbow Fluorescent Particles in that they contain a single fluorescent peak designed for checking the optical alignment of any flow cytometer. They also have low fluorescence and size coefficient of variation (CV). However, they contain similar fluorophores to the Ultra Rainbow Calibration Particles. As a result, they are used in every fluorescent channel of the flow cytometer from UV to IR. See Figure 6 for the histograms of Spherotech cat. No. URFP-30-5.
Image

Yellow Calibration Particles (YFPs)
The Yellow Calibration Particles (YCPs) are intended for the calibration of the FL1 channel of the flow cytometer. The MEFL values of the YCPs have also been determined. However, the MEFL value of the brightest peak (4.88 x 106) may be too bright for some users. In this case, the brightest peak can be put off scale to allow only four peaks to show on the screen.
The histograms obtained on a CyAn™ ADP analyzer and the MEFL value for all peaks are shown below in Figure 7.
Image

Allophycocyanin Calibration Particles (ACPs)
The Allophycocyanin Calibration Particles (Cat. No.: ACP-30-5K) contain a mixture of fluorescent particles with different intensities in the Allophycocyanin channel. They are excited by either a He-Ne laser at 632 nm or a diode laser at 635 nm. Similar to the other calibration particles, the MEAP values have been determined. As a result, the ACPs can be used to check the linearity of the instrument in the APC channel. Below in Figure 8 are the histograms obtained on a CyAn™ ADP analyzer and the MEAP value for all peaks.
Image

Blank Calibration Particles (BCPs)
The BCPs have similar background fluorescence to unstained cells. They are used to set the fluorescence threshold of the instrument. The BCPs are the dimmest peaks in the corresponding RCPs. Histograms of the BCP-30-5 (3 micron) are shown is Figure 9.
Image


Recommended Flow Cytometry QA Procedure
REAGENTS:
1. Sphero Blank Calibration Particles, Cat # BCP-30-5
2. Sphero Ultra Rainbow Fluorescent Particles, Cat. # URFP-30-2
3. Sphero Ultra Rainbow Calibration Particles, Cat. # URCP-38-2K
4. Sphero COMPtrol Compensation Particles, Cat. # CMIgP-50-3K

INSTRUCTIONS FOR USE:
Users are encouraged to modify the procedure described here to fit the QC needs of individual lab or personal preference.

A. Preparation of Particles
1. Vortex the particles vigorously.
2. Add 2-3 drops of calibration and fluorescent beads to 1 mL of sheath fluid. The inclusion of a small amount of
detergent (~0.01%) in the dilution buffer will increase the percentage of the singlet population which usually vary from
~75% to 90% depending upon the size of the particles, concentration of the particles and the dilution buffer. The unused
portion of the diluted particles suspension can be stored in the refrigerator for future use. If sterility is needed, the
particles can be washed once with 70% alcohol or 3% hydrogen peroxide by centrifugation and resuspension as follows:
a. Add 2-3 drops of particles to 1 mL of dilution buffer in 1.5 mL microfuge tube.
b. Centrifuge, remove the supernatant and resuspend the particles in 1 mL of 70 % alcohol or 3% hydrogen peroxide by
vortexing.
c. After 5 mins, centrifuge, remove the supernatant and resuspend in 1 mL of sterile dilution buffer.
d. Vortex, centrifuge, remove the supernatant and resuspend in 1 ml of sterile dilution buffer.
3. Vortex the diluted calibration and fluorescent particles briefly.


B. Daily Alignment
If available, use the single population particles such as RFP-30-5, RFP-30-5A, URFP-30-2, URFP-38-2, URFP-38-5A or other single intensity fluorescent particle to determine the quality of the optical system alignment. To determine the optical alignment of the system perform the following:
1. Set a live gate for the siglet population on the FSC vs SSC histogram to exclude aggregates.
2. Adjust the Gain and High voltage so that the mean channel number of the peak is in a predetermined position on each
histogram of interest. Fig. 5 or Fig. 6 can be used as a guide if the RFP-30-5 or URFP-30-2 are used.
3. Count a minimum of 5000 events inside the gate.
4. Record the CV, Gain, High Voltage and Relative Channel Number for FSC, SSC and all fluorescence channels of interest
as shown in Table 3. In addition, a computer program such as Excel can be used to generate the Levy Jennings graphs as shown Fig. 11 and Fig 12.
Image
Image
Image

If the values on any parameter exceed those of day-today average or preset values, which are determined by at least one months worth of data, additional calibration or alignment procedures should be performed according to the instrument operation manual.

If RFPs or URFPs are not available, align the optical system with RCPs and record all parameters of the instrument and the relative channel number of brightest peak or a designated peak in RCPs.


C. Setting the Threshold
Use the Blank Calibration Particles (Spherotech Cat. No. BCP-30-5) or unstained cells to adjust the forward scatter and to set the threshold. Place a live gate around the singlet population on the forward vs side scatter histogram.


D. Verification of Setting and Validation of Logarithmic Amplifier Linearity and Sensitivity
Use the RCPs or URCPs to verify the instrument settings and to check the sensitivity of the instrument.
1. Set a live gate for the siglet population on the FSC vs SSC histogram to exclude aggregates. Since these beads are
much smaller than blood cells, increase the FSC gain to place the beads on scale in the light scatter plot. Set a gate
around the singlet bead population. The Relative Channel Number of the initial dot display screen may look cluttered due
to the number of the populations and the aggregates. However, after setting a live gate on the FCS vs SSC, the dot
display screen is cleaned.
2. Set PMT voltages: Input the instrument settings normally used for specimens in your laboratory. In most instances, the
number of peaks will correspond to the histograms as shown in the package insert.
3. Turn off compensation.
4. Collect the plots for your panel, for example:
Forward Scatter-linear vs. Side Scatter-linear
FL1-log, FL2-log, FL3-log, FL4-log, FL5-log, FL6-log, FL7-log, FL8-log, FL9-log FL1-log vs FL2-log
5. Count a minimun of 5000 events inside the gate.
6. Record the peak value and channels of separation between adjacent peaks in a lab note book or on a computer program
as shown in Table 4.
Image
Other important data that should be collected include the Calculated Linear MEF, Slope, Intercept,
Percent Residual, and R2. See Spherotech STN-9 for information regarding these parameters. In addition, the
sensitivity can be determined and recorded by cross-calibrating the channel number of the blank particles against the
regression line created by the URCP-38-2K fluorescent peaks. The cross-calibration procedure is found in the
Spherotech STN-9. Collection of the recommended parameters will provide data to generate Levy Jennings graphs for
the long term performance tracking of the instrument. An example can be seen in Figure 13.
Image
These parameters should have values within the upper and lower cut off limit on a day-to-day basis. The upper and
lower cut off limits should be determined based on the collection of at least one months collection of data after the
instrument has recently beeen serviced. If the instruments performance does not fall within the cut off limits,
additional QC protocols or alignment should be performed according to the instrument operation manual.


E. Adjust Compensation
In order to correct for “spillover” or spectra overlapping of fluorescent molecule, compensation should be performed.. In order to use the Spherotech COMPtrol particles for compensation the following procedure has been developed.
1. The compensation tube must consist of particles that are unstained as well as particles that are stained with the
fluorescent probes of interest. The preparation of the COMPtrol particles is as follows:
a. Vortex COMPtrol Particles thoroughly before use.
b. Add 1 full drop (approximately 50μl) of each of the COMPtrol High Binding Capacity Particles , COMPtrol Low
Binding Capacity Particles and Negative Control Particles to a 1.5mL microfuge tube.
c. Add 10μl of the fluorescent monoclonal antibody conjugate (diluted to a concentration optimal for staining 10^6
cells) to the tube.
d. Vortex the tube well.
e. Incubate 15 – 30 minutes at room temperature. Protect from exposure to direct light.
f. Add 1-2mL of staining buffer to the tube.
g. Centrifuge the particles at 1500 x g for 1 minute and remove the supernatant.
h. Resuspend the particles in 500μl of staining buffer to the tube.
i. Vortex the tube well.
2. Set-up the flow cytometer PMT voltage settings using the target tissue for the experiment. The PMT voltages must be
set high enough to guarantee that the COMPtrol negative population is off of the axis in every channel.
3. Gate on the singlet bead population based on FSC and SSC characteristics. Set the analysis gate so that only the
COMPtrol particles are included.
4. Create a dot plot for the given fluorchromeconjugated antibody as appropriate.
5. Perform compensation procedures if necessary by aligning the centers of the COMPtrol particles populations by
matching the median fluorescences as shown in Figure 10.
Image