ATP determination kit: problem w/ final reagent preparation

PostPosted: Mon 21 Dec 2015 13:12
by blainejjda5
Hi,

I'm using the ATP kit and I have a little problem with it.
The kit works well when the final reagent is stored at +4°C but not when is aliquoted and stored at -20°C.
I used 2 kits (same batch) and obtained the same results: the fluorescence decreased rapidly in the week with the frozen aliquots.

Do you have an explanation?

Thanks

Re: ATP determination kit: problem w/ final reagent preparat

PostPosted: Mon 21 Dec 2015 14:05
by yeseniabolt
Hello,

it is very unusual that an aliquoted amount of the final reaction mix of the sensitive ATP assay losses activity when stored at -20 degrees and thawed once for use. Apparently, no mistake was made during your preparation of the final assay mix.
I advice you to not perform several freeze/thaw cycles with the assay mix, I tested up to 12 complete freeze/thaw cycles with a total loss of activity of < 20% compared to a fresh batch.
The loss of sensitivity/assay activity can also occur due to damage of luciferin by light. Even pipetting the assay in bright sunlight can have a negative influence of the assay sensitivity.
Also, Remind that using the sensitive assay it is strongly recommended to read out the samples after exactly 10-15 minutes. Longer incubation times with the reaction mix will result in weaker signals.

best regards