Rinse the external ELSD in a specific manner?
Posted: Tue 17 Mar 2015 10:05
Does the external ELSD need to be rinsed in a specific manner?
Yes, after each purification, the ELSD must be cleaned.
For this, purge the ELSD with isopropanol using the predefined "ELSD Nebulizer cleaning" method (this method takes 4 minutes).
You can stop the purge once the ELSD signal is stable.