How do I calibrate a puriFlash MS-HMW for ESI?

PostPosted: Fri 20 Mar 2015 08:45
by Interchim
How do I calibrate a puriFlash MS-HMW for ESI?

When calibrating a puriFlash MS-HMW for ESI, you should connect the source directly to the calibration source.
Prime gas introduction by pressurising the bottle, then stop the gas.
The calibration agent will continue to pass through the source by capillary action under the effect of the voltage.
This is sufficient for calibration and auto-tune, particularly for high masses.