Mass: Presence of calibration agent peaks, what should I do?

PostPosted: Fri 20 Mar 2015 08:48
by Interchim
Presence of calibration agent peaks, what should I do?

If you observe the presence of calibration agent peaks: to resolve this problem, rinse for 2h with acetonitrile and/or MeOH to eliminate all the calibration agent.
If you can still see traces, we recommend re-fitting the tube into the calibration agent bottle to ensure that it is no longer in contact with the liquid.
If the problem persists, there may be some calibration agent trapped in the system. You will probably need to replace a check valve and/or contact the After-sales service.