Pump leak detector problem (detector engages and pauses)

PostPosted: Thu 26 Mar 2015 17:12
by Interchim
Pump leak detector problem. The detector engages and pauses the instrument. The yellow detector LED stays on, what should I do?
Concerned instruments: puriFlash 215, puriFlash 430, puriFlash 450, puriFlash 4100, puriFlash 4250
Concerned module: pump


2 possible causes:
- Detector sensitivity too high
- A hydraulic tube (or its cable) passes near the end of the sensor


Adjust detector sensitivity by rotating the small screw in the hole next to the LED.
For adjustment, the LED must be off if there is nothing below it, and on when you pass your finger next to it.
If the problem persists, contact the After-sales service.