The solvent fumes are not extracted from the collector, what should I do?
Concerned instruments: puriFlash 215, puriFlash 430, puriFlash 450, puriFlash 4100, puriFlash 4250
Concerned module: fume enclosure
The exhaust tube is not correctly connected to the extractor, or the extractor is defective.
Check the tightness between the exhaust tube and the extractor, or replace the extractor.